They Don’t Make $37 Waveform Generators Like They Used To
[CreativeLab] bought a cheap arbitrary waveform generator and noted that it only had a two-pin power cord. That has its ups and downs. We feel certain the intent was to isolate the internal switching power supply to prevent ground loops through the scope probes or the USB connector. However, it is nice to have all your equipment referencing the same ground. [CreativeLab] agrees, so he decided to do something about it.
Opening the box revealed that there was hardly anything inside. The main board was behind the front panel. There was also the power supply and a USB board. Plus lots of empty space. Some argue the case is made too large to be deceptive, but we prefer to think it was to give you a generous front panel to use. Maybe.
It was a simple matter to ground everything to a new three-pin connector, but that left the problem of the USB port. Luckily, since it was already out on its own board, it was easy to wire in an isolator.
Honestly? We’d have hesitated to do this unless we had made absolutely sure it didn’t pose some safety hazard to “jump over” the switching power supply. They are often isolated for some reason. However, the likelihood is that it is just fine. What do you think? Let us know in the comments.
A similar unit had a reverse engineering project featured on Hackaday many years ago. While these used to be exotic gear, if you don’t mind some limitations, it is very easy to roll your own these days.
10 thoughts on “They Don’t Make $37 Waveform Generators Like They Used To
    1. Sometimes, too much. Which is why your microwave has an earth pin. On this world (signal gens, scopes, and other test gear), ideally none at DC, and controlled at signal frequencies.
      EArthing the signal ground is a many edged sword. You can float an unearthed output or input, but an unearthed output or input can float, which can reduce noise immunity.
      An earthed signal ground can lead to shorts to earth, bu with appropriate protection (GFCI), this can protect equipment from some typed of shorts.
      Many other edges here, including loading at frequency, max allowable ground float, and so on..
  1. First of all using blue and brown cables for earth wires is not a good idea, better to use a yellow/green wire to see that is an earth conductor. Now If it’s necessary yo have a ground reference, it’s better to have a ground post and ise it to have a ground reference.
    If the SMPS has a capacitive coupling between mains and the rest of the circuit and tis is a concern, maybe adding an insulting trasformer wit a electrostatic shiel between primary and secondary it’s a better idea.
  2. First, I consider this a quite bad video. @01:19 you get a glimpse of the power supply and there seems to be an RF suppression / filtering cap in the lower left corner of the power supply (Next to the transformer, there is even an isolation cutout under (or near) it.
    Hackaday’s concern about a “safety hazard to jump over the switching power supply is groundless . :) The Protective Earth terminal is meant to do exactly this: connect the metal parts of a case to GND.
    You don’t get a high quality power supply in a function generator in this price range, but apart from the overall quality of this thing, the leakage is normal. It’s not a safety concern for humans, as the current is very low, a handful of uA. This power supply does have input filtering and some electronics on the secondary side, I’ve seen much worse then this, but there is not enough info to judge whether that power supply is of decent enough quality to be safe. Chinese standards are not very high in this regard (Or they are not checked very thoroughly).
    When I sit on my chair and touch the tip of my scope probe, I can also see 100V on my Siglent scope. It’s just picked up from the mains wiring all over the house. This is all perfectly normal and harmless to humans, but I do agree that this leakage can damage electronics. You get a similar situation if the PE (GND) pin of your scope is not connected properly (Some of the mains sockets here in the EU accept plugs with PE, but the wall sockets do not have the PE). In that case you can feel a tingling when you touch the metal PC case. It can even be strong enough to hurt a bit, but as long as there are no real faults in the power supply, then it’s still just earth leakage from the input EMI filter and harmless (for humans).
    It’s also quite common to modify these function generators and put in a low frequency mains transformer with linear power supply. This has much lower leakage. Usually good enough to keep the whole box floating without an PE connection, and that has it’s advantages too. Always having your function generator grounded is not really ideal.
  3. Well, this reminds me a (not so) nice episode of my ancient life : The building where we worked was in two parts… each one with its own power circuit. The more recent was a standard 220V (yes french and EU standard). The other one was in fact on a 110 V power circuit, but “modernized” to 220 V using two 110 V lines in opposites phases… with the ground on the mid point.
    Now introduce some computer (a VAX 780), a bit of network, a remote Ethernet to serial server and some VT200 terminals. The VAX and the terminal server was on the ancient power line, but we needed to install some terminals in the new building. It was not too far (15-20 m), so we estimated that a well shielded RS232 serial line will be enough.
    After a terminal servers and two VT220 returned to Digital Equipment for being ineffective out of the box, we started to wonder if something else could be the cause.
    After a week of head scratching (and another terminal server + VT220 blown up), we discovered that the serial line shield was connected on both side, hence connecting two grounds with a 110 V between them.
    Since that time, I never look at two “ground” contacts the same way…
    1. This is why “Ground” and “PE” (Protective Earth) are two separate things. Normally the GND wire and the PE wire are connected to each other (On the utility side, before the ELCB (Earth Leagage Circuit Breaker). The shielding of the RS232 cable should never have been connected to “GND” in the first place, but only to PE.
      Also here in Europe (most) mains voltage plugs are not polarized. We do not really distinguish between “GND” and “Live” in single phase systems, but consider both wires “live”. This reduces the opportunity for mistakes between GND and PE by a lot. In houses the wires have defined colors Blue for GND, brown for live, black for “switching” and Green/Yellow for PE) but those wires are normally not visible by users. On a wall plug, you can not see which pin is “neutral” or “live”.
      Also, every now and then I see electronic schematics where the PE symbol is used as a generic GND for a circuit. Please don’t do that. Use the GND symbol instead for your low voltage uC circuit.
  4. i’m pretty sure i wouldn’t want this mod! My siglent oscope’s ground appears to be referenced to earth, and that isn’t a win for me. 90% of the time, it works out fine. But i guess sometimes i use un-isolated power supplies, and sometimes in order to make the most of two channel scope, I’m inclined to hook my scope’s “ground” probe to an in-between voltage instead of to the power-supply’s gound. Or even to two separate references for the two channels. And when i do that, the scope makes an awful noise and i disconnect it in a hurry and then i have to test whether i fried it. i think that’s happened 3x so far and each time it has not fried the scope and i have been able to figure out another way to make the measurement (or to live without it).
    i don’t really know the ups and downs of it. i halfway hope for someone to tell me there’s an option in the scopes menus to change this behavior! But i generally wish the scope was isolated. Being isolated is the one advantage i can point to of the old battery-powered scope that i used to have (and which was otherwise infuriating to use). If i had an isolated power supply that i wanted to ground-reference, i’d just run an external ground connection (jumper wire) so i can enable/disable this hack. Which is apparently basically what would happen if i happened to hook my scope up to any circuit coming off of this signal generator anyways.
    1. On nearly all scopes the BNC connectors are connected to the PE pin of the mains entry in a very similar way as the “hack” in this function generator. And this is done with thick wire too. Goal is that it can pass enough current to blow the mains fuses without damage to the scope. I’m mildly surprised it did not trigger your GFI (I guess the same as ELCB), but not everybody has those things. (But I do recommend to install one in the mains wiring to your electronics lab).
      And this is indeed a limitation of common scopes. If your scope does not state very clearly that it has floating inputs, then you can be pretty sure that the BNC’s are connected to PE (And I assume you have some electronic knowledge, so you can measure the resistance to the BNC shell and the PE pin).
      One of the possible remedies is to make sure that the gadget you are measuring is floating. This is one of the main uses of isolation transformers. Low voltage circuits are usually also floating relative to the mains voltage. Lab power supplies sometimes have an extra PE Banana / Screw connector near the output so you can choose whether the output is floating or mains PE / (GND) referenced.
      Another option is to use differential probes for your oscilloscope. In that case you don’t have to connect the scope “GND”. You can also attempt to use two scope channels and then subtract them from each other (a common scope function). This creates a “poor man’s differential probe”.
Leave a Reply to Suppressed CarrierCancel reply